the 120% and 80% of nominal concentration for accuracy study has to be exact or approximate concentration is okay?
I believe that's the *minimum* range over which you have to validate.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
the 120% and 80% of nominal concentration for accuracy study has to be exact or approximate concentration is okay?
I think you should target 80 and 120%. And would not need to be exact, we aren't. Usually 90-110% (or tighter) is the spec range for us anyhow for our products.