UV detector cell temperature

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello chromatographers,
I need to ask a question about temperature settings of the UV detector generally in HPLC systems, we previously had a Shimadzu LC 2010 CHT HPLC "1 block" sytem, and UV settings were either low, high or ambient, we used the default "low". Now we installed Shimadzu Prominence modular system with slight software changes, and cell temperature field requires a number input..
Can anyone tell me what is generally used and how to decide on the used temperature?
Temperature can influence both the absorption coefficient and the wavelength max. For that reason it's not a bad idea to control it. So, I think it's a nice feater you have there.

In most cases and probably in yours as well, it is adequate to set the flow cell temperature at the same value as the column temperatur.

Best Regards
Learn Innovate and Share

Dancho Dikov
2 posts Page 1 of 1

Who is online

In total there are 347 users online :: 0 registered, 0 hidden and 347 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 347 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry