Hey Adam, I agree with much of what you say, with respect to the sensitivity / flow rate / injection volume issue. Scale the injection volume, and there is no sensitivity advantage. I am actually glad to hear someone else say this.
I guess I am often sample limited. Not to the extent that the proteomics guys are, but I might want to inject a sample a few times to get statistics on it, or I might want to reinject a check standard many times out of the same vial rather than having multiple vials. I can inject 4.7 times less volume with a 2.1mm column versus with a 4.6mm, and get the same sensitivity.
On the dead volume issue, again I feel your pain. I've seen LC setups that ship with mass specs, that are plumbed with too much dead volume to use with 2mm columns. I run 0.005'' i.d. tubing everywhere between the pump and detector, when possible.
If someone came to me and told me that I should run 1mm or 0.32mm columns, I would give many of the same reasons you gave for not doing it, so I guess it is a matter of degree. But, if I had an older ion source that was very inefficient at desolvating the droplets, and a 2mm column required a splitter, then I might scale down to 1mm.
Which leads me back to the splitter issue and clogging, where I must disagree. I will assume for the sake of discussion that we are using ordinary splitters, not active splitters. The mass spec sees the same thing when there's no clog. When a clog starts to develop, the two situations are different. As a clog begins to form, there will be a pressure drop across the point of the clog. As the flow at that point becomes more and more restricted, the pressure drop continues to increase, up to some maximum. In the split-flow situation, the maximum pressure drop experienced, even with complete plugging, will simply be the backpressure from your waste line (perhaps a few hundred PSI at most). In the non-split situation, the pressure drop at the clog point will increase much more rapidly as it approaches total plugging, up to the maximum pressure specified in your method (perhaps 4000-5000 psi). In this situation, the clog is either going to be short-lived, or it's going to shut your system down.
Now imagine all the times you've ever had an ESI needle clog. Some of them could be "blown-out" with maybe 1000psi or less from your pump LC pump. I'll call these "minor" clogs. Some were indeed bad enough that even 5000psi could not blow them out (time replace the needle or some tubing). I'll call these "major" clogs.
In the non-split situation, only the major clogs will shut you down. The minor clogs will either persist without shutting you down, or they'll be "blown-out" during the course of your run and be transient events, only causing a problem if they happen at the wrong time. And, a persistent minor clog will not change the flow rate that your source sees for more than a second or two.
In the split-flow situation, even minor clogs will shut you down, or at the very least change the flow rate that your source is seeing. And if a partial clog causes your source to see 20uL/min instead of 200uL/min, you won't have any immediate warning of it.
For qualitative ID, I think splitters are fine, and sometimes will use one rather than scaling down an existing method. For quantitation, I am leery of using them.