Optichrom Peak issue
Posted: Mon Mar 14, 2011 7:39 pm
The setup
GC: Optichrom Advance
Service: EP streams analyzing Methane, Carbon Dioxide, Ethane, Propane, C4+
Detector: 30KΩ TCD
Carrier: Helium
Sample valve and column valve are both 6 port Model 11
Reference flow 22 CM3/M
Carrier flow ( Det vent) 40 CM3/M
Reverse Column Step
Columns:
1 5ft 20% 0V-101 – 80/100 CHROM P
2A 1 ft HAYESEP Q -80/100
2B 2ft HAYESEP P – 80/100
2C 14 ft CHROMOSORB 750 – 80/100
3 1 ft 20% 0V-101 – 80/100 CHROM P
With CV1 on flow is Carrier >1 > ITC > 2 >3 > DET > Det Vent
With CV1 off flow is Carrier > 2>ITC<1>3 > DET>DET vent
My issue with this analyzer is both the CO2 and the Methane Peaks are reading much higher than they should be on one of the streams. The CO2 is reading anywhere from 0.01% off to 0.04% off. The Methane is reading about 2.5% or 2.75% off.
I tried checking that stream on the other two streams on this analyzer (just used a jumper on the tubing over to the other stream) and the results were pretty consistent on all streams. This leads me to think that it is either a sample issue or a timing issue on the gates and reverse column step. I replaced the vaporizing regulator at the sample point which is a long distance away from the analyzer. I am not sure if the regulator is working properly as there is no documentation on them that I have been able to find, they were rebuilt in 09 so they were not new in the warehouse. Any documentation that may have been with them new is long gone. Go regulator PN H2-108358 which I have been told is a custom part for Siemens?
I can upload screen captures of the grams if needed.
thoughts on where I should go from here?
Thanks
GC: Optichrom Advance
Service: EP streams analyzing Methane, Carbon Dioxide, Ethane, Propane, C4+
Detector: 30KΩ TCD
Carrier: Helium
Sample valve and column valve are both 6 port Model 11
Reference flow 22 CM3/M
Carrier flow ( Det vent) 40 CM3/M
Reverse Column Step
Columns:
1 5ft 20% 0V-101 – 80/100 CHROM P
2A 1 ft HAYESEP Q -80/100
2B 2ft HAYESEP P – 80/100
2C 14 ft CHROMOSORB 750 – 80/100
3 1 ft 20% 0V-101 – 80/100 CHROM P
With CV1 on flow is Carrier >1 > ITC > 2 >3 > DET > Det Vent
With CV1 off flow is Carrier > 2>ITC<1>3 > DET>DET vent
My issue with this analyzer is both the CO2 and the Methane Peaks are reading much higher than they should be on one of the streams. The CO2 is reading anywhere from 0.01% off to 0.04% off. The Methane is reading about 2.5% or 2.75% off.
I tried checking that stream on the other two streams on this analyzer (just used a jumper on the tubing over to the other stream) and the results were pretty consistent on all streams. This leads me to think that it is either a sample issue or a timing issue on the gates and reverse column step. I replaced the vaporizing regulator at the sample point which is a long distance away from the analyzer. I am not sure if the regulator is working properly as there is no documentation on them that I have been able to find, they were rebuilt in 09 so they were not new in the warehouse. Any documentation that may have been with them new is long gone. Go regulator PN H2-108358 which I have been told is a custom part for Siemens?
I can upload screen captures of the grams if needed.
thoughts on where I should go from here?
Thanks