I'm an intern and for my graduation paper I have to determine a UV-stabilizer (HALS). They gave me the tip of detecting it with an ELSD, but nobody in the lab is familiar with the detector. Please understand that I'm kinda new in all of this, I'm graduating from university this year.
I found out which peak is the UV-stabilizer, it elutes at +- 35 minutes but it's very wide and the baseline is very noisy.
I'm using a Waters 2424 ELSD, the column of the HPLC is a C18 of 15cm.
the drift tube is on 60°C
the nebulizer is on cooling
gain 3
The Hostavin N30 is solved in tetrahydrofuran with concentrations of 10, 100 and 1000 ppm.
I use a mobile fase gradient:
A: milli q water + 0.5mg/l n-hexylamine
B: acetonitrile + 0.5mg/l n-hexylamine
C: isopropanol + 0.5mg/l n-hexylamine
According to the journal I've been following, the hexylamin is to block some C18 groups. Otherwise the stabilizer wouldn't elute.
time - A - B - C
0 - 30 - 70 - 0
2 - 30 - 70 - 0
34 - 0 - 100 - 0
36 - 0 - 0 - 100
41 - 30 - 70 - 0
45 - 30 - 70 - 0
Does somebody have any tips on getting a better baseline and why it is that the peak of the stabilizer is so wide?
thank is advance!