I recently developed a method in which four components of a mixture were resolved. The system I used was 0.5mM NH3 and 0.24 mM TFA and the pH was near 9.5 or so. All my sample components come off the column within 35 minutes.
The column is an Alltech Novosep A-1 Anion which utilizes tetra alkyl ammonium.
The problem I am having is the peaks become very broad when I increas the sample load and I lose resolution.
I believe the NH3 is important to keep the pH above 8 and keep my sample ionized. I think TFA is my ion exchanger.
I was wondering what I can do to increase sample load without losing resolution and retention. Previously I increased my buffer concentration and each component eluted faster and I lost resolution. For example at 6 mM TFA and 12 mM NH3 everything in my sample elutes in ten minutes where at above mentioned conditions they come out within 35 min.
No organic was used. All aqueous.