MX304 » Tue Feb 22, 2011 7:48 pm
The HED supply is located on the board that is on the side of the door, correct?
I have replaced the board with another one that I know works and it still has the same error.
The HED supply is in the electronics box below the door. Follow the "spark plug wire" for lack of a better term from the side board down. That wire is connected between the side board and electronics board. Before you get too carried away swapping electronics, or calling Agilent let's try a few easy things that I have seen cause this over the years. First, make sure that HED lead is fully connected at both ends. If you have another instrument, try swapping that lead and see if the problem goes away. Take the multiplier horn out and clean the contacts. I have seen multipliers loose good contact and give this error. Make 100% sure all the leads going to the source are in the correct position and making good contact. I have seen this error several times, but only once did I end up replacing any major parts.