Hi,

I would like to know whether there is any requirement that internal standard need to elute at the end (i.e. after the analytes elute) during an analysis?

Thanks for your time.
valk

No requirement, but often considered "better", particularly with MS detection because it minimizes the possibility that something will change the ionization efficiency after the IS and before the analyte. My personal opinion is that the IS should elute as close as possible to the analyte; whether just before or just after is not that big an issue.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

I would add that what is most important is that the internal standard accomplish the purpose for which it is added. Using an internal standard is not "free". the price paid is an inherent decrease in precision (because two areas are used in the calculation) and a doubling of the chances for coelution in complex samples. So there needs to be some demonstrated justification for using the internal standard. In most cases where I have seen an internal standard used in LC it didn't add to precision and I can quote cases where it made precision greatly worse. If you are using the internal standard to compensate for detector response drift, demonstrate that the compound chosen responds to detector changes like the analyte. Ditto for what ever other purpose an internal standard is used.

It is very easy to calculate results from data files with and without internal standard calculation. Do it both ways for a while to prove that the internal standard is doing something useful. You may find that it is not.
Bill Tindall

That´s the key, do it both ways, it doesnt cause any extra work (if you start out with the Int. St. method anyway), . . . maybe a little extra calcs. With the Int. St. there, you will at least have an extra assurance that things are going oK, during real analyses.
On the calcs, any book on chrom, or any integration software instructions will tell you.

adding to Bill's post -- and if you find the internal standard is not helping (or is making things worse, then that gives you some insight into the source of error (in your procedure after the IS and analyte have begun to separate).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 5 users online :: 3 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot], Semrush [Bot] and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry