In an FID detector the hydrogen is burned in the flame. So as long as the flame is lit, there is no hydrogen escaping into the room. If the flame is out and the gas flow is on, hydrogen is released. (Modern instruments will turn of hydrogen if the flame is not sustained.) In labs I have been in, hydrogen diffuses so quickly and the air turnover in the lab is sufficient that there is no need for concern over accumulation of hazardous levels of hydrogen from the FID or septum purge - even on older instruments.
There was a time that one would set flows with a flow meter, and after getting them all adjusted and gasses turned on, one would press the igniter button. There would be a pop and the flame would be lit. Ocasionally the flame would not light - which was determined by lack of signal and lack of fogging a mirror (all of which would take several minutes to check - with gas flowing) and the next step would be to push the ignitor button to see if the flame would light and stay lit. Pressing the button would produce a pop from down in the instrument - not a blast, knocking small tools from the top of the GC or such. The gas stream leaving the FID mixed with the air in the room so quickly that the only hydrogen igniting was that contained in the FID or very close by the exit from the FID.
If you place an exhaust vent over the FID,you are introducing an air current that has the potential to affect the flame. I would suggest against that unless you are working with compunds that are so toxic that you would have a concern about what residual material might survive the flame and go free in the room.
The split vent, which may deliver considerably more hydrogen into the room (depending on settings), a different story, which Rod addressed above.