we r having TCD
TCD 175C
Column 200C
Injector 150C
i tried many conditions also but not found the better one.
plz help me........................
Regards
N
Advertisement
Discussions about GC and other "gas phase" separation techniques.
we r analysing mix of acid, alcohol (etoh, buoh) with some quantity of water and aldehydes using Porapak Q column. Resolution of alcohols water is ok but acid (acetic acid n butyric acid) was not good.
we r having TCD
TCD 175C
Column 200C
Injector 150C
i tried many conditions also but not found the better one.
plz help me........................
Regards
N
And either of those will separate all those at once isothermally? I don't have a packed inlet, but maybe it's worth seeing if I can get one.Column must be glass or fused silica lined SS, 80/100 Carbopack C/0.1% SP-1000 3 meters 80°C, maybe higher.
Or
80/100 Carbopack B/5% SP-1000 2 meters, maybe longer 90°C-110°C range, pick one.
That lets that column out.If you were to pick a capillary, the of course, SPB-1000 would be the column of choice, megabore as thick a film as is available, but IPA and EtOH separation would be minimal or none.
In the past I've used a PoraPlot-Q capillary ... seemed to shed a lot of particles (FID spikes)... in any case i can't delete the acids.If you delete the acids, then Porapak Q, 2m in fused silica SS tubing with screens, but only buy from Supelco unless you want voids in the packing that will ruin the separations. ( I no longer work for Supelco)
That's what would be best if it's possible.Getting acids as reasonable peaks is a problem if you want to separate them WITH the alcohols at the same time, isothermal and under 10 minutes SINGLE DIMENSIONALLY.
I've used a DB-624 as well as DB-Wax for the alcohols, but from that experience i doubt the acids would elute isothermally in any reasonable timeframe on those columns.Your best bet is the reliable OV1301 phase capillary (aka DB-624 SPB-624) as the issue of ethanol, acetone, and IPA are well separated on a few columns but tail on many, and the water degrades the capillary.
For the foreseeable future probably between 60 - 100/day.... Which is why sample prep also has to be fast/simple.How long, how many samples, etc will you be expecting to do this analysis?
I was thinking of using headspace...I know from a different assay in a different life, from a much longer temperature programed run, I can get reasonable results for acids (R=0.998+ and RSDs < about 3%) using HS at these concentration levels, if I control pH. Running the HS oven at 60-70C would not put that much water on the column. The main issue is finding a column that will let me do all the analytes isothermally in 10 minutes or less.Free acids DO NOT do well on most columns, and add water to the mix, well, it can get VERY expensive.
Thanks. Sounds like if I figure it out, it would be publishable.Good luck, but as I have said before, if it were easy.................................
In the past they were linear doing that.A Caveat
I anticipate that unless your adjustment of pH is done with a strong and non-volatile acid acid your recoveries of the acids will be non-linear.
Thanks... It all hinges on finding the right column... and I do have an idea on something to try.And I do wish you the best of luck,