In my paper in Analytical Chemistry my regression line for extremely small concentrations for 18 solvents including toluene, from a mixture of water and DMAc, were all > 0.9995 except for very basic pyridine which was 0.995. That was at concentrations of 0.5, 10, 25, 50, 137, 484, and 1000 ng per sample vial. (nanogram is correct, it is not microgram)
These were not exceptional results, but results that were gathered during routine work period between regular daily samples. I know another researcher has done much better than this; right, Peter A. in Africa?
When preparing stds one should STANDARDIZE as much as possible. Putting the spiked additions into five potions of the same sample solution is one way to minimize errors and to demonstrate the true accuracy of the method.
You have good reproducibility from all appearances and I believe you will demonstrate even better results.
I have always found HS to be extremely precise and it would actually show my errors in preparation when properly executed.
best wishes,
Rod
I was focussing almost exclusively on repeatability rather than linearity (this was in the SA National Metrology lab with a very specific focus of being able to compare between samples). If I recall correctly headspace in other applications and other labs could give r-squared of 0.999 if everyting was well tuned (but this was usually at higher levels than Rod was using). As a last resort I would sometimes skirt around repeatability by running duplicates or triplicates.
1% rsd is actually better than I would expect from an off-the shelf headspacer (I never tried a Perkin Elmer) so the high recovery is not a repeatability issue.
How certain are you that the unspiked material has 100 ppm (not a proper unit by the way, I've been nagging another poster about it !) ?
By implication you are adding different volumes of toluene solution to give you different spiked concentrations (and presumably that volume was zero for the unspiked samples). If the diluent for the toluene has a matrix effect in the samples this might account for both the high recovery and the somewhat poor linearity. Best practise is to add a constant volume of a spike solution whose concentration is varied to give the different spike concentrations.
Peter