This week in our lab:
Our "formulation scientists" were playing with different presevatives in a granules formulation. In one batch we found 130 % of one preservative (lets call it P2). We repeated the analysis, reviewed the batch record .... no results. When I checked the identity spectra the PPI was around 970 in both standard and sample. When I checked the spectra there was a slight difference at our detection WL. When I subtracted the standard from the sample spectrum I got a distinct difference spectrum with a maximum at our detection wavelength. That was exactly the spectrum of another preservative (P1).
Result of checking the log of the granulator:
They made some batches granules with P1, cleaned it and made for the beloved analysts a placebo batch containing P1 (doesn't make much sense to make a placebo with out all that junk that is in the stability testing batches, but even less sense is to have a placebo with API in).
As we are detecting P2 at one of the valleys in the UV spectrum and P1 has at this WL its maximium (around 4 AU at concentrations employed) traces of P1 are vastly overdetected.
Basic line: the Peak purity index of standard was 973 and the peak purity index of the sample was 970, so no difference, however 30% impurity at the detection WL, coeluting.
I have to measure the preservatives at their absorption minima to get the in the same scale as the API that is two magnitudes lower in concentration.
Thats on peak purity testing. Monday I will spend on the OOS procedure.
Alex