The peak area response trend is random ie. no general decrease within vial or between vial for the same (std or sample) solution. Therefore, degradation is not suspected.
1. An internal std compound eluted later in the chromatogram shows very good RSD, hence ruling out instrument issue. Also, high %RSD (>5%) was observed for multiple inst.
2. The response generally decrease with multiple injection from the same vial (60% methanol:40%water as sample solvent & mobile phase) but there are some exceptions where the %RSD is very good.
3. Tried different LC columns, insts, LC vial/caps, etc.....still high %RSD.
4. Diazolidinyl urea exist a complex mixture (apparently 2 peaks) but the UV spectra of the peaks are very similar therefore the response should be similar.
5. UV spectrum is very steep at the selected wavelength (for optimal response) but I don't suspect it will cause high % RSD within a run.
6. I don't believe it's related to the compd pKa.....does anyone know it's pKa. The sample solvent is appprox pH 8.4. Could it be the mp is close to the compd's pKa? I plan to use a pH 5 buffer solution to check.
I tried HILIC but I'm not getting the chromatography outlined in a literature paper. The diazolidinyl urea compound elutes at the solvent from for all the different columns/mp combinations tried.