Method sensitivity issue- Please help
Posted: Sun Feb 27, 2011 10:35 pm
by rgonsal
I'm developing a method for bisphosphonates using Ion-pair RP (tetrahexyl salt). I'm already limited with the sample concentration of 0.05mg/ml. I need sensitivity @ 50 nanograms/ml. The problem i'm facing is due to the tailing of this bisphosphonate, the peak area is good enough (but the height is small). Injection volume is 200-250 microliter on 250 x 3.0 mm, 5micron column. detection = 215 (210 is max). How can I inject a larger volume without dispersing the sample band or increase the signal?
Suggestions are very appreciated.
Thanks!
Re: Method sensitivity issue- Please help
Posted: Mon Feb 28, 2011 2:28 am
by tom jupille
To get "on-column concentration", the diluent should be much weaker than your mobile phase. You didn't specify either one, but 250 microliters on a 3mm ID column is already getting fairly large (the equivalent of over 500 microliters on a 4.6 mm ID column), so I don't know how much more you could get.
LOD and LOQ are ultimately limited by signal/noise ratio. If you can decrease your baseline noise by 10x, that drops your LOQ by 10X. So:
- if you move down toward the 210 nm absorbance maximum, does your signal increase more than your noise?
- how much noise do you have (and what is the manufacturer's spec for noise on your system)?
- have you tried to optimize detector settings (things like bandpass, time constant, etc.)? If you're using a PDA, you can play with these ex post facto to see what happens.
- is your pump operating at it's best (degassing, check valve operation, etc.)?
Every little bit helps.
Posted: Mon Feb 28, 2011 2:37 am
by rgonsal
Hi Tom,
The diluent is the mobile phase. The IP reagent itself has absorbance @ 215/210 nm (changing to 210nm didn't help). The detector used is a PDA , but seems like nothing works. I have to explore the variable VWD (we have Agilent's LC in our lab) in the hopes of having a lower detector noise. I have been using Optima grade ACN as well.
How can I possibly eliminate the tailing of bisphosphonate inorder to get a sharp peak?
Posted: Mon Feb 28, 2011 6:41 pm
by tom jupille
If your diluent is the mobile phase, then 250 microliters is a tremendous volume overload on a 3mm ID column, and I suspect that's the genesis of your peak shape problem. What *is* your mobile phase? and can you use something weaker (i.e., less organic solvent) as the diluent?
What is the noise spec for your detector?
How much noise are you actually seeing?
What, exactly, have you tried that hasn't worked?
Posted: Tue Mar 01, 2011 3:33 am
by rgonsal
Hi Tom,
I don't know the exact noise spec for the detector. The mobile phase is phosphate buffer with EDTA pH=7.70 (900ml) combined with 100ml of ACN.
Posted: Tue Mar 01, 2011 4:45 am
by tom jupille
You might gain a bit by using water or buffer as your diluent (your mobile phase is 10% ACN).
You should be able to find it in the manual or on the manufacturer's web site. If you are operating in a regulated environment (GLP or cGMP), you should have documentation on the performance qualification (PQ) tests run on your system. That qualification should have included a noise measurement.
Once you know what the detector *should* do, you can measure what it *is* doing to see if improvement is possible.