- Posts: 33
- Joined: Mon May 25, 2009 5:53 am
is there anyone out there who has had much experience with using manual SPME with a merlin microseal?
After a few injections I have been getting problems with reproducibility (large sample loss), the 'duckbill' on the seal seems to abrade quickly and there are sometimes CO2 ion fragments in my mass spec which make me think I had an air leak at the time of injection. (Not to mention the suspicious 'pish' sound that it makes when I put the needle in and out.)
I use an SPME liner and inject in splitless mode. The SPME needle is left in the inlet after the purge vent opens for the duration of the 12 min run to condition it. The SPME needle is blunt-tip, 23 gauge.
I am currently trying samples using a standard septum which seems to be working fine and just wondering whether anyone else has had problems with the merlin?
cheers
Liz