I have a Varian CP 3800 GC with a type 1 EFC (1079 injection port).

The total flowrate of carrier gas is around 500 cc/min and I cannot find a way to reduce it. Also the pressure in the injection port is too high. The pressure at the gas cylinder is well within its limits.

I can reduce the flowrate to around 120 cc/min, but only by reducing the carrier gas pressure and then pressure in the injection port is too low.

I believe the problem is a broken flow controller but I was hoping someone might have some experience with a similar problem.


Thanks in advance.

Hi Sean,
This is a vague memory from 5 years ago:
I had a similar problem when we had a CP8400 autosampler strapped onto the CP3800 - 1079 EPC. I think a different circuit board was required and subsequently it needed recalibration (?) since the carrier and split flows increased drastically.
Perhaps you have a circuit board failure?

WK
Hello Sean,
check also the splitline, often here You can find condensate or dust or something else.
Check also the connections between the pneumatic-electronic part and the top of the instrument (Viton inlets 1/16" tubing)

Regards
Bernd
bm

Hello Sean,

Do you disassembly the left cover of the 3800 gc; perhaps you've catch splitline, or like Bernd sayed before the split line is blocked by condensate or dust.

Regards

On this topic of 3800:
Have you had to replace either the 3800 carrier trap or 3800 split trap?
Does either affect flow?

On my experience i never change the split trap and carrier trap
But do we speak about the same think trap under the 3800 GC, and not external trap before GC ?

BG - thanks for the info - I did mean the traps inside the 3800.
Sean - sorry for extending the thread.
The internal plumbing of the 3800 is a little mysterious and full of connections - not helped by a lack of diagrams in manual.

Sean,

I really think the pb is due to an obtruction of a 1079' tube.
Clean your injector by heating in first time to see the result and change eventually your insert (or control the insert mounting !)
8 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: uzman and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry