Oops! Haha sorry about that. I've never really done anything like this before.
The instrument used was a Varian GC.
The column is a Restek RTX-5 30m x 0.25 mm x 0.25 µm.
Helium carrier gas was used for the analysis.
The samples were manually injected by hot needle injection.
The a few of the retention times were labelled on the chromatogram, but for the majority of the samples, the retention times were calculated from the graph by hand. The same was done for the peak widths, calculated by hand.
I hope that's enough information, but I doubt it since the calculations for the peak widths I have done are on the chromatograms. Later today Ill head to the library and scan them.
EDIT:
Didn't make it to the library, but I found a way to get the pictures of the chromatograms.
The test mixture that was used for the analysis contained n-C13, n-C14 and n-C15. The compounds were dissolved in carbon disulphide solvent. The results below refer to the isothermal runs with increasing linear velocity of 12, 20, 30, 50 and 70 cm/s.
http://img825.imageshack.us/i/isotherma ... arvel.jpg/
http://img600.imageshack.us/i/isotherma ... arvel.jpg/
http://img827.imageshack.us/i/isotherma ... arvel.jpg/
http://img571.imageshack.us/i/isotherma ... arvel.jpg/
http://img716.imageshack.us/i/isotherma ... arvel.jpg/
These images are closeups of the n-C14 peak which was needed to plot a Van Deemter curve.
http://img9.imageshack.us/i/isothermalr ... eupof.jpg/
http://img812.imageshack.us/i/isotherma ... eupof.jpg/
http://img101.imageshack.us/i/isotherma ... eupof.jpg/
http://img841.imageshack.us/i/isotherma ... eupof.jpg/
http://img46.imageshack.us/i/isothermal ... eupof.jpg/
I just want to make sure I've got the right idea in mind to create the plot.
So first, the air peak time is calculated using the Peterson and Hirsch method. The retention times can be found directly from the chromatogram which is the time from injection to elution.
Second, the peak width is measured at either the baseline or at the half-height of the peak (I used both in excel to validate the answer) using a ruler. I used ratios here to calculate the peak width (For example, if the scale was 2.2 cm = 0.1 min and the width I measured is 3.4 cm, then I would put it as a ratio and solve getting 0.1545 minutes).
The number of theoretical plates can then be found by using the peak width and retention time.
Next, the plate height is found by using the column height (30m = 30000mm) and the number of theoretical plates.
I was asked to plot height vs. the actual linear flow rate for the C14 peak, so I would find the average linear velocity first? Then change it to linear flow rate?
EDIT 2:
Ok so I changed up my units to make them more simpler, but my hopes to that accomplishment would enable me to find the reason why my graph looks quite different from most graphs. I played with the values to see how it would affect my graph and it turns out that if I change the theoretical plate height for sample 4, the graph resembles much like a Van Deemter plot. I was suspecting that the problem was hidden with either sample 4 or 5 and it turns out that sample 4 is a possibility. I'm now trying to understand how one could get to this kind of conclusion. For the plate height to be affected that way, the number of theoretical plates has to change and so does the retention time and the peak width. Right now Im looking into what could affect the retention time and peak width.
Heres my edited data: