Concentration of buffer of mobile phase in RPC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
12 posts
Page 1 of 1
Does concentration of buffer of mobile phase have an effect on retention time in reversed phase chromatography,for eg; If I change my buffer concentration from 0.3 mM to 0.6 mM ,will it have any effect on retention time for the peak of interest in RPC?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Most of the time, so long as there is sufficient buffer to control ionization, the actual concentration above that has little effect on retention. Typical buffer concentrations for LC with UV detection are in the 10 - 50 mM range. At below 1 mM, you may be below the "suficient" level.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
ESK - you're the researcher - research this !!! This is the type of thing done in robustness studies, like how far can the buffer concentration vary without affecting the results !!!Does concentration of buffer of mobile phase have an effect on retention time in reversed phase chromatography,for eg; If I change my buffer concentration from 0.3 mM to 0.6 mM ,will it have any effect on retention time for the peak of interest in RPC?
-
- Posts: 20
- Joined: Fri Dec 04, 2009 5:54 pm
I did a study on this with some basic drugs since the silica plays a role in retention. At 2.5 mM buffer concentration k' was over 10, at 50 mM k' was 4. Tailing inmproved significantly as well.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
The pH was the same?
-
- Posts: 20
- Joined: Fri Dec 04, 2009 5:54 pm
Yes, it was a low pH phosphate buffer.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I'm not surprised, but it would have been interesting to see intermediate results. Based on my experience, I would expect to see a big improvement from 2.5 to 5 to 10 and then leveling off. The results would depend quite a bit on the on the silanol activity as well. Newer ("type B", "high purity", . . . etc.) packings can tolerate lower buffer concentrations. Regardless, if the concentration is below 1 mM, I expect that it *will* matter.I did a study on this with some basic drugs since the silica plays a role in retention. At 2.5 mM buffer concentration k' was over 10, at 50 mM k' was 4. Tailing improved significantly as well
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
What is "low pH"? I am still much surprised at such a large ionic strength effect if the pH was low enough to eliminate SiO-.
-
- Posts: 20
- Joined: Fri Dec 04, 2009 5:54 pm
pH was 2.5, done on an older silica column. Yes, both the tailing and the effect on k leveled off. Not much improvement above 20 mM for the compounds I tested.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Do you have organic modifyer in your mobile phase? How did you determine pH? I still think this is a pH effect, if it is really an ionic strength effect I would be strongly interested of the identiy of your compounds.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
pH might be low enough in principle, but if the buffer concentration is too low (hence low buffer capacity) the residual silanols will still cause trouble – especially (but not exclusively) in the case of the silica A support.
Best Regards
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 345
- Joined: Sun Dec 06, 2009 6:16 pm
Monitoring the effect of buffer concentration on retention and peak shape is the classic experiment to determine the extent to which ion exchange plays a role in the separation.
Tom's advice is sound regarding the typical minimum buffer concentration (5-10mM) to mask the ion exchange effects that will be observable with basic compounds on many of the newest packings out there.
Tom's advice is sound regarding the typical minimum buffer concentration (5-10mM) to mask the ion exchange effects that will be observable with basic compounds on many of the newest packings out there.
A. Carl Sanchez
12 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science