Same column, conditions, different GC =retention time shift?
Discussions about GC and other "gas phase" separation techniques.
16 posts
Page 1 of 2
Been experiencing a pretty significant and consistent retention time shift of a volatile analyte when toggling between two Agilent GC's. The column is the same column, same inlet liner, same operating conditions, but on one instrument the analyte has a retention time of 3 minutes and on the other GC the peak of interest has retention times of 4.8 to 5 minutes. Thought the differences may be due to a leak somewhere, but this has been going on for months, with many column installations and consumables being replaced. Leak test shows no leak in inlet, gold seal, etc. Split vent trap and split vent line replaced. I have checked all column dimensions entered. Any ideas of what could be responsible for this consistent retention time shift?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
"Once upon a time .................. and the driver said, officer, I was only going 55."
Just because you set a number into a keypad connected to a computing device controlling a piece of hardware, doesn't mean the hardware is actually doing what you set it to do.
Check flows. Check temperature. {Both by a hand-held, third party device.}
[ I assume the GCs are in the same lab, at the same distance above sea level. ]
When you assume something to be factual, and you don't tell me the 'facts' are assumptions, then
it is possible to make an ASS of U + ME.
Never ASSUME.
best wishes,
Rod
Just because you set a number into a keypad connected to a computing device controlling a piece of hardware, doesn't mean the hardware is actually doing what you set it to do.
Check flows. Check temperature. {Both by a hand-held, third party device.}
[ I assume the GCs are in the same lab, at the same distance above sea level. ]

When you assume something to be factual, and you don't tell me the 'facts' are assumptions, then
it is possible to make an ASS of U + ME.
Never ASSUME.
best wishes,
Rod
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
I'll second what Rod says - verify all the operating conditions.
This exact problem came up on this forum in the not so distant past, try searching the archives.
And it is difficult to offer good trouble shooting advice when you do not tell us what the analyte is, the column(s) or any of the operating conditions - the only hard facts you give are the manufacturer of the instruments and the two retention times.
Peter
This exact problem came up on this forum in the not so distant past, try searching the archives.
And it is difficult to offer good trouble shooting advice when you do not tell us what the analyte is, the column(s) or any of the operating conditions - the only hard facts you give are the manufacturer of the instruments and the two retention times.
Peter
Peter Apps
-
- Posts: 6
- Joined: Wed Jan 26, 2011 9:29 am
Are you sure its the GC conditions and not because one column has degraded.
Have you tried swapping columns between kits?
Have you tried swapping columns between kits?
-
- Posts: 170
- Joined: Thu Mar 30, 2006 9:35 pm
same column
same inlet liner
same operating conditions
toggling between two Agilent GC's.
Also, what analyte?consistent retention time shift
Two different instruments will not always perform identically.
John
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Stupid question same gc method? Is one constant flow while the other is constant pressure.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
And have you checked the calibration of the inlets for readign pressure exactly. If yoy cannot do that measure the dead time of the column (methane with a warm GC oven and inlet - or if you do not have natural gas available, you can use butane from a cigarett ligher - if you can find one, but use a failry hot GC oven.) and see if an unretained compund moves through the instruments at the same speed.
-
- Posts: 30
- Joined: Sat Sep 25, 2010 5:17 pm
When I checked the flow using butane, the unretained compound had a retention time almost double what it should be for a 15M .530mm column. With the GC set to a linear velocity of 24, the actual average linear velocity was found to be 13.5 cm/sec. This explains the almost doubling of retention times. Could this large of a discrepancy be corrected by simply zeroing the EPC module?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
First check to be sure that the column configuration inclulding detector is correct for the column. It is not that hard to accidentaly turn vacuum correction off or on in the wrong place. And be sure that columns one and two are set to different inlets. Off the top of my head this seems to be a fairly good sized jump for a re-zero. (I've seen both columns set to one inlet accidentally - and it can really mess things up, even though the settigns for the one column actually installed i nthe instrument are correct.)
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
Also check that you have set the correct carrier gas in the GC control software - different gasses need different pressures to get them flowing at the same velocity.
Peter
Peter
Peter Apps
-
- Posts: 30
- Joined: Sat Sep 25, 2010 5:17 pm
Thanks for everyone's input. I had some time to check some configurations and still can't pinpoint the cause of the excessive retention times for the analyte (EtOAc) / unretained compound. Using the same column and inlet liner on 3 different instruments - Agilent 5890, 6890, and 7890... It is the 6890 that is experiencing the excessive retention times. I went over the methods on all 3 instruments in hopes of finding an operating condition I may have overlooked. So far I've checked:
leaks (none using a pressure test), column dimensions, correct carrier gas selected (He), correct configuration of inlets/detectors, vacuum correction off, constant flow on all methods, same temperature program. Changed the split vent trap and split line.
The unretained compound, butane, has a retention time of almost double on the 6890 as compared to the 5890 and 7890. On the 6890, linear velocity was set at 24 and was calculated to be 13.5.
Could there be a delay in integration, and if so where can I check for that in Chemstation?
Could there be a valve of some sort that is getting partially stuck upon injecton?
I have not fooled around with recalibrating the EPC because I am unfamiliar with this process. Is this something that I should look into next? What are the caveats when re-zeroing EPC? I'm scratching my head on this one so any insight is much appreciated.
leaks (none using a pressure test), column dimensions, correct carrier gas selected (He), correct configuration of inlets/detectors, vacuum correction off, constant flow on all methods, same temperature program. Changed the split vent trap and split line.
The unretained compound, butane, has a retention time of almost double on the 6890 as compared to the 5890 and 7890. On the 6890, linear velocity was set at 24 and was calculated to be 13.5.
Could there be a delay in integration, and if so where can I check for that in Chemstation?
Could there be a valve of some sort that is getting partially stuck upon injecton?
I have not fooled around with recalibrating the EPC because I am unfamiliar with this process. Is this something that I should look into next? What are the caveats when re-zeroing EPC? I'm scratching my head on this one so any insight is much appreciated.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Note the pressure reading on each GC at start run time. If they show as being the same, get a pressure gauge and measure the head pressure. (They make - or at least used to make pressure guages for GC's).
An other thing to help sort this all out. The 6890 should gie you the linear velocity of the column. Since you know the length of the column, you can compute the elution time of the butane, assuming that it unretained. My guess is that one GC is close to the expected retention time (assuming a warm GC oven to keep from significantly retaining the butane) and the other will be off. This will point a finger at a GC.
An other thing to help sort this all out. The 6890 should gie you the linear velocity of the column. Since you know the length of the column, you can compute the elution time of the butane, assuming that it unretained. My guess is that one GC is close to the expected retention time (assuming a warm GC oven to keep from significantly retaining the butane) and the other will be off. This will point a finger at a GC.
-
- Posts: 30
- Joined: Sat Sep 25, 2010 5:17 pm
Don,
I did calculate the linear velocity for the 6890 and it was 13 whereas the gc display read 24. The question is why...
I did calculate the linear velocity for the 6890 and it was 13 whereas the gc display read 24. The question is why...
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
Just to be absolutely sure - when you say the same column, same inlet liner etc etc do you mean that you have put the same physical column into each instrument, or that each instrument has its own column, and that they are the same length, diameter and stationary phase ?
What type of inlet iner(s) are you using. ?
Peter
What type of inlet iner(s) are you using. ?
Peter
Peter Apps
-
- Posts: 30
- Joined: Sat Sep 25, 2010 5:17 pm
The same physical column. The liner is agilent 4mm with glass wool...#5183-4647 and is used in all 3 instruments.
16 posts
Page 1 of 2
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science