Problem with inertsil SP column
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
For one perticular molecule containing phosphorus, I am using INERTSIL SP column. The mobile phase contains tetrahexyl ammonium sulfate, ETDA, TEA, Phosphate buffer(pH=7.5) and ACN. The method is isocratic. But after few injections,the peaks become broad. There is no regeneration procedure available on the net. Can anyone help?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
What you are seeing is consistent with garbage from your sample accumulating on the column. There is no "universal" regeneration procedure (and there is no guarantee that you can ever recover the column). In general, what you need to do is to reverse-flush your column with something that is a good solvent for whatever kind of garbage is there (obviously, that depends on the nature of the sample). Things to be watch out for:
- Make sure that any solvent change doesn't involve precipitation of buffer or miscibility problems.
- Disconnect the detector before flushing the column.
- If possible, flush the column in the reverse-flow direction (check with the vendor first !!).
Moving forward, you will need to either improve the sample cleanup or use a guard cartridge (and change it frequently).
There is also a chance that the problem is coming from contaminated mobile phase. It is possible to tell this apart from sample contamination, but the procedure requires some patience. There is a description on our web site here: http://www.lcresources.com/resources/TSWiz/hs390.htm
- Make sure that any solvent change doesn't involve precipitation of buffer or miscibility problems.
- Disconnect the detector before flushing the column.
- If possible, flush the column in the reverse-flow direction (check with the vendor first !!).
Moving forward, you will need to either improve the sample cleanup or use a guard cartridge (and change it frequently).
There is also a chance that the problem is coming from contaminated mobile phase. It is possible to tell this apart from sample contamination, but the procedure requires some patience. There is a description on our web site here: http://www.lcresources.com/resources/TSWiz/hs390.htm
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science