Dear all, I am eagerly seeking for advice..

Initially, I have set up an UPLC method for peptide analysis using an Acquity BEH C18 column.
FYI:
Mobile phase A: 90% 25mM KH2PO4 (pH 7.4), 10% MeOH, 2 mM ammonium acetate, 0.1% f.a
Mobile phase B: 10% 25mM KH2PO4 (pH 7.4), 90% MeOH, , 2 mM ammonium acetate, 0.1% f.a
gradient conditions: 90%A (0-6 min), 50%A (6-9min), 90%A (9-12 min)
flow rate:0.5 mL/min
Tcolumn:25C
Tsample:10±5C
Peptide eluted at 4.594 min

Then, I got split peaks. Hence, I performed a thorough cleaning of the UPLC system and column following Acquity restoration guidelines. Split peaks were removed.

With the aim of getting a shorter run time than the one of 12 min, I changed MeOH to AcN
To avoid buffer precipitation, mobile phases were changed to:
Mobile phase A: 90% 25mM KH2PO4 (pH 7.4), 10% AcN, 2 mM ammonium acetate, 0.1% f.a
Mobile phase B: 100% AcN, 2 mM ammonium acetate, 0.1% f.a
Tcolumn:40C
I tried a number of isocratic conditions (60%A, 50%A, 30%A) and flow rates from 0.2-0.5mL/min but peptide eluted very early (max retention time of 0.6min at 0.2mL/min)
I tried a number of gradient conditions and flow rates of 0.6mL/min (since increasing flow rate would decrease run time) but peptide eluted similarly very early. Even when I performed the "MeOH gradient" described above (note different mobile phase conditions,Tcolumn and flow rate) peptide eluted at 0.3 min.

I am in a mess really.Any feedback would be appreciated.
Regards,
T
Hey there -

Chromatography of peptides, even fairly small ones, is different from other small molecules. Under reversed phase conditions, isocratic HPLC of peptides is frequently not possible. Below a certain organic concentration, you'll have essentially infinite retention and above it you'll have none. I must admit that I do not entirely understand your motivation for having ammonium acetate, formic acid, and potassium phosphate in you mobile phase, either.

I would suggest a gradient between an aqueous phase containing a perfluorinated carboxylic acid ion pairing agent at low pH and either acetonitrile and methanol.

I don't have much time now to explain further - meeting - but will get back later on.

Cheers!
CJ
http://the-ghetto-chromatographer.blogspot.com/
Here I am..
Well, my peptide is highly recommended to be in phosphate buffer conditions (pH 7.4). I have already thought of moving to sodium phosphate instead (especially after reading quite a few tips - LCResources). I am using ammonium acetate and formic acid as mobile phase modifiers since I have observed that they improve peptide chromatography (in general) and help out sample ionisation when there is the potential of coupling UPLC/HPLC methods to MS (as in my case).

Regards
T
Looks like you have gone from a system where the buffer level remains the same regardless of mix ratio to one where the level of buffering varies with %B present.

the Acetonitrile may be having an appreciable effect on the pH of your system which may acount for the retention problem.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
any thoughts?? shall i go back to MeOH?(I went to AcN in order to gain better peak widths) or shall I go for a %B of lets say (50%AcN, 50% phosphate)? would you also think that this is not due to column "age" ? my mind is in such a mess!
5 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry