I'm setting up an LC-MS method for "B", which is an aromatic thing approx 200Da with heterocyclic rings, no bonds of purely single-bond character, and no obvious leaving groups. My instrument is an ion trap. Not surprisingly, "B" doesn't fragment at reasonable energies, and is detected dreadfully at very high energies.
Question 1: is it OK for me to quantify a "fragement" ion that is actually the precursor? I've never done this before, but because the molecule is so tough, doing SRM of M -> M is actually adding selectivity and removing noise. I have about 6 peaks in full-scan MS2 of the precursor ion in real samples, but if I select the parent mass and use an energy at which everything else fragments, I have only one peak remaining, with good S/N.
Question 2 is about internal standard calibration. My client has added an internal standard "A" which is closely related to salicylate, and which I think was originally put in the method because the method started out as an assay for salicylate. I recover about 60% of A (external standard calibration) from all his samples. He spiked two of his samples with B, of which I recover about 50% (by external standard calibration). My client naturally assumed that he could scale up my reported amounts of B by 100/60, and get quite a happy result.
The problem is that when I do internal standard calibration "properly", with a set of standard containing increasing B and constant A, plotting cali-curve of area ratio versus amount of B, the answer for recovery of spiked "B" is still only 50%. The reason is that in the standards, when I add a lot of B, the signal for A starts to fall. It can't be cosuppression because the two chemicals do not coelute. I suspect it must be something chemical going on. On the other hand, the spiked samples show the same signal for A as all the others, rather than showing an extra-low signal for A. Therefore the effect is either not additive, or the real samples contain something that stops B from removing A. (Or something very strange happened and the two samples that were spiked just happened to have a much better recovery of A before they were spiked... highly unlikely!). At the level my client chose to spike, the signal of A has already fallen to about 60%, so the calibration curve assigns the 60% signal of "A" to being caused entirely by the presence of "B", and concludes there was no loss during extraction.
My feeling is that I can't trust internal standard calibration using "A" on these samples for measuring "B". I don't really believe that there is no loss during extraction, or no loss of efficiency of measuring A in real samples (based on what I see in all the un-spiked samples). Since the calibration curve for B tries to behave otherwise, I don't really believe it's working. The bottom line is that I'm concerned B isn't sufficiently chemically similar to A.
I'd appreciate any thoughts on (1) what might be going on, chemically/physically, and (2) am I right to abandon this internal standard? What would you do, faced with this sort of situation? I'm a bit puzzled.