lmh » Tue Feb 22, 2011 10:26 am
I'll also repeat what Tom said!
Yes, it would be even less logical to use peak area ratios if the product peak has exceeded the reasonable range of a calibration curve, but the whole process is, in any case, illogical. But there is no alternative.
In typical LC-MS ionisation (e.g. electrospray) the ionisation efficiency (signal per mole) of different chemicals can differ wildly, even when the chemicals don't actually appear all that different. Ideally we'd like to know the percentage impurity expressed in some useful basis (percentage by mass, molar percent, etc.). Because we have no idea of the ionisation efficiency of the impurity (because we don't have it as a standard) we can't convert area percent to molar or mass percent (irrespective of what we know about the impurity's mass). The area percent is fairly meaningless, because a strongly-ionising impurity will give a much larger value than a weakly ionising impurity. Unfortunately, what else can we do? "No standard" means "no calibration".