carryover in LCMS
Posted: Mon Feb 21, 2011 2:59 pm
I use an Agilent 1100 HPLC coupled to AB Sciex 4000 QTrap to analyze some samples for multi-analytes (approx. 16) in the sequence: Blank (50% mobile phase), AQC (standards spiked 50% mobile phase), NQC (blank urine), same blank urine sample spiked with one analyte, positive QC (urine spiked with all analytes).
Unexpectedly, an analyte not spiked was detected in the urine sample in addition to the spiked analyte. 2 MRM transitions were positive and in high count (AQC: 4e6, urine sample: 3e6). The compound identity was further confirmed by Enhanced Product Ion full mass spectrum.
I did not set needle wash. But there was no carryover detected at a concentration of 5000 ng/mL in the validation experiments before (the concentration was greater than the analyte's spiked concentration in the AQC (100 ng/mL).
No analyte was detected in the NQC sample immediately after the AQC.
I wonder if carryover can occur to the urine sample even after an NQC in between.
Please help.