Chromeleon - removing smoothing filter
Posted: Thu Mar 03, 2005 3:05 pm
by dblux
Could somebody of Chromeleon 6.60 users let me know how to remove smoothing effect (a form of additional channel for sample) from chromatogram data file ? TIA
Wojtek
BTW - Chromeleon manual describes clearly how to apply this filter, but there is no word on how remove it. It's not a "manipulation" so "delete manipulations" command doesn't work in this case.
Posted: Thu Mar 03, 2005 6:41 pm
by Mark Tracy
First of all, you need privileges both in CM and Win. You may need to ask your CM system admin. Be sure that this does not violate some company policy.
1. Unprotect your data source: In CM, right-click on the data source, select properties, and uncheck "Protect data directory." Close the dialog, exit CM and log out of Windows.
2. Remove the data: Log back in to Windows. From Explorer, find the folders with the smoothed data and drag those folders to desktop. Their names will be the names of the synthetic data channels. Start CM and confirm your work. Delete the folders.
3. Protect your data source: As in step one, check "Protect data directory." Log out of Windows and log back in.
Be careful,
Posted: Thu Mar 03, 2005 6:57 pm
by dblux
Thank you for your very professional reply. As I can see this operation is not for an average Chromeleon user and even one of service engineers didn't know this procedure. Thx again.
Wojtek
Posted: Thu Mar 10, 2005 5:33 pm
by Mark Tracy
I have just received information from our Chromeleon support people on a much better and safer way to to this:
1. click on the sequence (not the sample line inside the sequence)
2. press Delete key; a dialog box appears with several options
3. select the option "Only selected raw data" and a list of data channels will appear.
4. check the smoothed channel for deletion
5. click YES
I apologize for the awkward approach I suggested earlier.
Posted: Fri Mar 11, 2005 7:52 am
by Andreas Gr
Dear dblux,
Thanks for pointing out the missing hint in the help. We entered it as #13296 to our Tracker system and added it to the Chromeleon help.
Andreas Grömping
Project Manager Chromeleon Documentation
Dionex Softron GmbH
