After speaking with Varian tech support I found out that my system could not see ions below 40 m/z despite numerous efforts by me and Varian. I ended up cleaning the ion trap as a last resort this past Friday and letting my system bakeout over the weekend.
This monday I found that my system could now properly see ions across the entire size range (10-650). But now the detector is seeing maxed out intensity across the entire range!
I was under the assumption that if my turbopump was properly reaching 100% speed then I did not have a leak, but Varian support told me that I could still have a leak. They also seem convinced that an air leak is causing the maxed intensities on my spectrum. I have spent two days now checking every o-ring and seal that I can find but I cannot locate the leak.
I read about a popular leak-detecting method involving spraying compressed air around the possible leak locations but this does not work as my intensities are all maxed out so I cannot see the peak for the air propellant.
Any suggestions for possible leak locations or other solutions would be greatly appreciated.
Thank you
edit: I do not think the problem/leak could be in my GC because before cleaning the ion trap the system was seeing all ions fine (above 40 m/z) and I did not make any alteration to the GC while cleaning the ion trap.