by
winsun » Wed Feb 23, 2011 7:37 am
First, you could run this product for solvents by headspace. Since it is a liquid that boils at a high temperature, it should not be a problem to put 50 or 100 microliters into a headspace vial and run it neat with 3 std additions in DMF or DMSO (don't forget to add blank DMF or DMSO to your 0 ppm spike)
IT is unlikely you have much ethanol present since the boiling points are SO different it should be easy to separate EtOH and DEM. Unless of course, the DEM is degraded from the presence of water and over time.
You could also run this test as a v/v test injecting directly onto a column such as a porous polymer column.
Methanol, ethanol, n-propanol should all elute before DEM. If you do it v/v use a suitable late eluting solvent for your stds (toluene comes to mind).
I would run the column isothermally at 180°C, with injector at 200°C and detector at 250°C
Headspace could be done isothermally at 80°C
best wishes,
Rod
Hi Rod, really thank you very much for detailed explanation and information.
To be frank, I am sales manager of our company, so I only know little about the test, actually we don't have to have this kind of test for our DEM and DEMM products before, one of our overseas client is now requesting this from me, so I am seeking some information on internet and from experts like you.

Actually I also found something with very detailed information like Chromatography conditions, Sample preparation, Operation procedure and System adaptability, but our client said they followed that method, but failed, so if you could give me some information or revision to the below method, that will be greatly appreciated. Daniel
HEADSPACE GAS CHROMATOGRAPHY TEST FOR RESIDULE ETHANOL
Chromatography conditions:
Chromatography Column:DB624,30m×0.53mm×3um or 30m×0.32mm×1.8um
Carrier gas: N2 or He
Flow speed: 35cm/s
Split Ratio: 1:5
Headspace temperature: sample vial temperature 80℃, sample loop temperature 105℃, transfer line temperature 105℃
Sample conditions: heating balance 45min,sampler vial pressurization 0.5min,sampling 1.0min
oven temperature:40℃ keep 20min, 20℃/minheating up to 240℃,keep 20min (may reduce properly)
Detector: FID,temperature 250℃,inlet temperature 140℃。
Sample preparation:
Sample solution: Accurately weigh 0.5g sample,use DMF to dissolve and dilute to 50ml
Solvent solution: Accurately weigh 0.5g ethanol,use water to dilute to 100ml,Accurately weigh 1.0ml diluted solution,use water to dilute to 100ml
Blank solvent:use Solvent solution method to prepare blank solvent,but without ethanol
Sample:Accurately weigh 5.0ml sample solution+1.0ml blank solution (3 parallel samples)
Reference sample A:Accurately weight 5.0ml sample solution+1.0ml solvent solution (3 parallel samples)
Reference sample B:Accurately weight 5.0ml DMF+1.0ml blank solution
Operation procedure:
—Reference sample B inlet
—Reference sample A inlet
—Sample inlet
System adaptability:
Reference sample A’s 3 syringes peak area relative standard deviation must be lower than 5.0%