Setting total flows for Agilent 5890 split inlet with EPC
Posted: Sun Feb 13, 2011 5:50 pm
I recently changed the column in my Agilent 5890 (split inlet with EPC and MS detector).
Now the max flow through the split vent I can get is only about 10 mL/min.
(I was able to get up to 100 mL/min plus with the previous column set up.)
I am wondering the best approach to trouble shoot this problem.
I don't seem to have an air leak since there is no evidence of that in the MS spectrum
so I assume the column is properly installed in the inlet.
The column I removed was a MolSieve, 30m, 0.32mm
The new column is a DB1, 40m, 0.18mm. I've set the head pressure to 25psi at 100deg C.
Any suggestions to approach this problem are appreciated.