by
stan83 » Tue Feb 15, 2011 8:59 am
Thanks Rod,
I checked my TCD, Perkin Elmer is saying that there is no problem when working with a capilary columns 0.32 and 0.53 mm without any problems and without aditional make up gas if the flow is bigger than 5cc/min. These are the specifications of the TCD:
Operating 100 °C to 350 °C in 1 °C increments temperature.
Sensitivity with a detector temperature of 100 °C : 9 μV/ppm nonane at 160 mA at the bridge
Minimum detectable quantity: Typically < 1 ppm nonane
Linearity: > 105
Power supply: Constant current with four selectable settings:
1: ±40 mA
2: ±80 mA
3: ±120 mA
4: ±160 mA
Signal filtration: 50, 200, 800 msec
Filament protection : Self-limiting and resetting after transient overloads in either channel
Makeup gas: Not required for 0.32- to 0.53-mm i.d.
columns with flows ³ 5 mL/min
Required for 0.25-mm or smaller
i.d. columns
So I guess it should not be a problem. However, the problem here arrises when I want to analyse CO2 as well. I have a packed Molsieve 5A column, but I cant let CO2 pass to it. So the current method is letting only the permanet gasses (CO, N2, O2 and methane) to pass through PoraPack Q and to Molsieve. Next eluting component is CO2 and the second column, Molsieve is closed untill all other components (CO2, ethane, ethylene, acetylene) come out from the PoraPack, after which Molsieve is again connected.
What I was reffering was the brochure for the Micro-GC by agilend :
http://www.youngin.com/application/AN-0809-0098EN.pdf
they managed to separate permanent gasses on molsieve capilary column and light hydrocarbons + CO2 on a poraPlot Q in less than 1 minute each. However they have a 250 nanoliter TCD

. Therefore I want to use same length of the columns and similar flow, however I do not expect so fast analysis and so low detection limits.