I've found HFBA to be useful for separating synthetic peptides with Arg deletions. Most of the time TFA is sufficient to resolve Arg deletions, but there were instances when TFA system could not separate the deletions peptides no matter how shallow the gradient. In these cases, HFBA did the trick. More recently I've used mixture of 0.08% TFA + 0.02% HFBA, and this was also effective.
I think someone mentioned- i've experienced baseline fluctuations with HFBA. I have even used freshly distilled stuff without much better results. This is a problem when monitoring below 220 nm at high-sensitivity detector attenuation.. Depending on the sample (Y, W content), i've had luck monitoring at 280 nm, where, Unlike at 220 nm, the attenuation can be cranked all the way up without a great deal of baseline noise.
If this prep hplc, consider if hfba counterion is ok for subsequent application