karthic » Mon Feb 08, 2010 3:02 am
I use this column too. Initially, I had difficulty in separating oxygen and nitrogen but after developing a proper method, it has never been an issue. Argon is the right carrier gas.
Follow this method:
Oven
Inital temperature at 32 C for 12 min (this long wait is to see teh separation between oxygen and nitrogen) then ramp at 30 C/min till it reaches at 236 C to remove CO, methane and CO2.
Well, again the procedure would depend on what dimensions your column is. I m using a 30 m long column with a nominal diameter of 320 um and a film thickness of 15 um.
Also, I have a ramped flow rate. It is initially kept at 0.4 ml/min for 12 min and then ramped at 0.1 ml/min2 till the flow reaches 0.4 ml/min.
Also, I use very small sample size (100 ul) and have a split ratio of 30:1.
Hope this helps you. Feel free to ask questions if you have any doubts.
Karthic