I need to determine purity of a high purity commercial chemical (N,N'-diethyl-N,N'-diphenylurea - molecular mass: 268; boiling point: 330 ºC). I'm using the mass balance approach (100 % - total impurities) to do it.
I'm using GC-FID (DB-5 30x25x0,25 - split 1:50 and 1:25 - inlet and detector temp 250 ºC) to get chromatographic purity. I can see 3 impurities.
Impurity 1: Retention Time 6.8 min
Impurity 2: Retention Time 7.1 min
Impurity 3: Retention Time 32.9 min
N,N'-diethyl-N,N'-diphenylurea: Retention Time 33.3 min
When I went to GC-MS to try to identify the impurities by making use of the library it came out:
Impurity 1: ethylbenzene
Impurity 2: p-xylene
Impurity 3: non identified by the library, but molecular peak was 254 (it seems to be related to N,N'-diethyl-N,N'-diphenylurea - very close retention times - and similar molecular mass - impurity 3 and main compound molecular mass has a diffence of 14 which indicates a CH2 absence in the impurity, so I guess impurity 3 is N-ethyl,N'-methyl-N,N'-diphenylurea. Furthermore, the fragmentation of this compound fits the real GC fragmentation). N-ethyl,N'-methyl-N,N'-diphenylurea boiling point is 364.6.
N,N'-diethyl-N,N'-diphenylurea DSC purity experiment came out 99.83 mol% purity (very high purity).
The impurities are present in a very small concentration. This is why I prepared 7 mg/mL solutions for the GC-FID analysis and the injection volume was 1 microliter (split ratio 1:50 or 1:25). I need to come with high concentration to see the impurities.
The problem is: After the first run, I I ran a solvent blank (methanol), and a little bit of N,N'-diethyl-N,N'-diphenylurea (retention time 33,3 min) appears, like a ghost peak. Even though the peak is so small compared to the previous run, after a second run of the real sample, the small peak would contribute to the peak area of this new run. As I need to make very accurate measurements, I have to get rid of this very small amount of N,N'-diethyl-N,N'-diphenylurea from previous run.
The small peak disappears after 1-2 solvent blank runs or if I condition the inlet using a very high temperatyre (350-370 ºC).
What are the suggestions to get rid of the "ghost" peak? It seems N,N'-diethyl-N,N'-diphenylurea doesn't volatilize completly and part of it remains in the inlet and goes out in the next 2 runs.
I use an inlet temperature of 250 º C. Should I try to raise it? Won't it cause problems to the other peaks, like ethylbenzene and p-xylene. At the very end, I'll do peak area normalization and the peaks have to represent the real amount of each one in my real sample.
I appreciate your contributions.
Best regards,
Rodrigo