Changing Chromatogram

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello! I have a problem. I am performing simultaneous quantification of vitamins A, E and D and last week I was able to separate all the peaks. The working conditions were:
mobil phase hexane:isopropanol (99:1), 1ml/min, 290nm. The column is Nucleodur NH2.
This week I tried again and now ganma tocopherol overlap with Vitamin D. I have repited sevaral times and it's the same. I have equilibrated the column for 2 hours before the firs running and I have made fresh solutions. Does anyone knows what is the problem?

Thank you in advanced
hi, how you prepare vitamin E stock solution from viscous standard.
Is the overlap due to wider peaks or to a change in retention?

If it was a change in retention, did all your peaks shift or just one of these?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Wow, I haven't used "normal" phase in three decades. But I remember that the trace water picked up by solvents could be important to causing irreproducibility.
4 posts Page 1 of 1

Who is online

In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry