Overloading a column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I was wondering, if you realized that the column appears to be overloaded, is there any good cleaning advice that could do a good job of reviving the column? I'm using a Zorbax 2.1 x 150 mm, 5 um column.
I consider "overloading" as too much sample in an injection, like beyond the capacity of the column packing, resulting in very ugly, rounded peaks.

If you feel your column needs cleaning or regeneration your Zorbax (assume RP-18, there's many different types of Zorbax) likely can be helped by high levels of organic, like 90% high levels of ACN or methanol. For more-involved cleaning, some run a series of IPA, then CH2Cl2, then IPA again, and then back to mobile phase. Most modern columns can also be reverse-flushed.

Where you need to start is with the operating manual for your column, or to call Agilent tech service 800-227-9770.
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry