How can we determine the flow rate of the carrier gas and sample? In my case I have to determine hydrogen gas conc. in a gas sample which contains unknown conc. of hydrogen.
-
- Posts: 6
- Joined: Fri Feb 04, 2011 4:22 pm
What is the best detector for Hydrogen gas conc. determination?
Thanks
Thanks
-
- Posts: 6
- Joined: Fri Feb 04, 2011 4:22 pm
Whats d best carrier gas for Hydrogen gas conc. determination. I have options like Argon, Helium and Nitrogen. Which one suits best? I have read that Nitrogen is the best carrier gas. Does it apply to detect hydrogen gas determination?
Thanks
Thanks
-
- Posts: 6
- Joined: Fri Feb 04, 2011 4:22 pm
What happens to the sample after it gets separated in the column? How does the detector sense the sample (or) the actual connection between the column and detector?
Plz respond soon.
Thanks
Plz respond soon.
Thanks
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I would recommend that you get hold of a good book on basic GC. McNair's book would be a good choice. Here is a link to it on Amazon:
http://tinyurl.com/4mx8rml
http://tinyurl.com/4mx8rml
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
I am jumping to a conclusion but from YOUR words, you wanted the flow rate of your sample and of hydrogen. So I assume you are measuring the output of a mixer, mixing a gas sample and hydrogen together and you want to know their concentration in the mixed sample.
Nitrogen is the better choice to measure hydrogen compared to argon or helium.
The thermal conductivity detector measures the changes in cooling rate of a hot wire as different gases pass over it. The gas with the greatest difference in conductivity to hydrogen is the better choice.
You will have a near linear response over two or three orders of magnitude if you use a detector of good design, the proper sample size, and the proper flow rate. These have to be experimentally determined, not quoted as a melting point or density value.
The sample that is injected onto a column is hopefully separated into its components on the column and hopefully all component gases will elute off the column given enough time. Some gases like water may not elute and may stay trapped on the column.
A little study as Tom suggested is a great idea. Perhaps looking at Gow-Mac's web site and other resources on the web might also prove useful.
I am certain you can perform a web search on the topics you seek answers, so I won't do your work for you on that level.
Good luck in chromatography,
Rod
Nitrogen is the better choice to measure hydrogen compared to argon or helium.
The thermal conductivity detector measures the changes in cooling rate of a hot wire as different gases pass over it. The gas with the greatest difference in conductivity to hydrogen is the better choice.
You will have a near linear response over two or three orders of magnitude if you use a detector of good design, the proper sample size, and the proper flow rate. These have to be experimentally determined, not quoted as a melting point or density value.
The sample that is injected onto a column is hopefully separated into its components on the column and hopefully all component gases will elute off the column given enough time. Some gases like water may not elute and may stay trapped on the column.
A little study as Tom suggested is a great idea. Perhaps looking at Gow-Mac's web site and other resources on the web might also prove useful.
I am certain you can perform a web search on the topics you seek answers, so I won't do your work for you on that level.
Good luck in chromatography,
Rod
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science