AKTA 2 pump systems
Posted: Mon Feb 07, 2011 10:40 pm
Hi, I am somewhat of an FPLC novice. I am currently writing a separation method for a customized AKTApurifier (2 pump system: system pump and sample pump).
When writing methods based on CV or volume, setting the pump flow to zero holds the program indefinitely. I can get around this by writing separate blocks for buffer pumping and sample loading. However, I believe in one of the thousands of manual pages it recommends setting the system (buffer) pump to a low flow rate (e.g. 0.1mL/min) while sample loading. I think the continuous flow probably prevents contamination to some degree. Aside from this, does anybody see any benefit to pumping both pumps simultaneously? I suspect that the slight flow of buffer is responsible for deviations from AKTAxpress (1 shared pump) results.
Is there anything I'm overlooking? Thanks for any light that can be shed.
Andy
When writing methods based on CV or volume, setting the pump flow to zero holds the program indefinitely. I can get around this by writing separate blocks for buffer pumping and sample loading. However, I believe in one of the thousands of manual pages it recommends setting the system (buffer) pump to a low flow rate (e.g. 0.1mL/min) while sample loading. I think the continuous flow probably prevents contamination to some degree. Aside from this, does anybody see any benefit to pumping both pumps simultaneously? I suspect that the slight flow of buffer is responsible for deviations from AKTAxpress (1 shared pump) results.