dear all,

ihave problem in complete samples up to volume , ihave sample contain 3 active ingredient

and 2 of them pass the limit of acceptance criteria but the last one failed with result of 94.7%

and the acceptance criteria is 95% . all of these active were in the same dilution the same pippet

my diluent methanol&water 1:1.

may ihave answer for this mistakes.
Without knowing any of the details, I can only speculate:

The third compound was impure?

You weighed incorrectly?

The third compound is unstable in solution?

The third compound is only marginally soluble, and some precipitated before the dilution?

The third compound adsorbs onto glass surfaces?

The third compound peak was integrated incorrectly?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Is acceptance criteria 95% or 95.0%?

94.7% rounds to 95%, amount of significant figures is, well, significant...
3 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry