In GC temperature gradient design, the initial oven temperature should be below the boiling point of the injection solvent.
This is not true in all cases, and having the solvent condense onto the column can cause some quite dramatic peak distortions
I wonder what the solvent in BSTFA + 1% TMCS (Supplier: W R Grace) is. I can't find its solvent in the reagent insert or from the internet.
Surely it must be on the MSDS ?
Any expert knows the answer?
Please help.
Solvent condensation will cause problems if there is enough of it to form a lens across the column that then gets pushed along by the carrier gas. In a 0.25 mm diameter column this can happen with less than 0.2 ul. The problem is worse if there is a polarity mis-match between the stationary phase and solvent because then the solvent has less tendency to spread and more tendency to form lenses. Avoiding these solvent-stationary phase interactions is one reason to use a retention gap.Thanks.
"having the solvent condense onto the column can cause some quite dramatic peak distortions"
Peter, I thought the solvent can condense on the liner and cause solvent effect. Could you please elaborate more on the situation when solvent condensation is not desirable?
Peter and Don_Hilton, I will look up the MSDS for the content. Thanks.