below is a quote from the empower manual :
Injector Connections
Detectors and system controllers that are not controlled by an Empower system (for example, 2690/2695, 600 Series controller, and GC) require an inject start trigger signal (contact closure) from the injector as each injection occurs. The inject start trigger signal instructs detectors and system controllers to acquire data or run methods.
Waters autosamplers transmit the inject start signal over the IEEE-488 bus or by trigger wire. If you connect a Waters 2690/2695, 717, or 717plus to an IEEE-488 interface, you do not need to make any additional connections for an inject start signal.
When under IEEE-488 control, the 2690/2695 Sample Management System, 717, and 717plus Autosamplers transmit the inject start signal directly over the IEEE-488 bus during data acquisition. All controlled devices are triggered simultaneously when an injection is made by these instruments. Trigger wires are not required when all instruments in a chromatographic system are controlled over the IEEE-488 bus and the busSAT/IN Module.
Note: Any instrument that is not controlled by Empower requires an inject start trigger wire connection.