- Posts: 3
- Joined: Wed Feb 16, 2011 4:57 pm
After pump down, when the source and quad heater temperatures are moving to the setpoints, the source heater reading was highly volatile, ranging from ~180 to ~520. Chasing the error numbers in Chemstation lead to the diagnosis of 'Source RTD not connected'. The quad temperature readback varied slightly, but not as extremely.
I should mention that after the first error, I pulled the source and cleaned it and made sure the connections were from the heater to the board were solid. No improvement was found on pumpdown.
Not wanting to spring for a full source heater assembley from Agilent, I ordered just the source temperature sensor from sisweb (the sensor shown here: http://www.sisweb.com/ms/hp/5973heater-sensor.htm). That arrived today. It pumps down now and the performance is better in that I'm not getting any error messages.
I am worried that this was a bandaid at best. The source temperature now fluctuates ±10 C from the setpoint. Example - setpoint 230, heats to ~240, then cools to ~220, heats to ~240 ad nasuem. The quad temperature is rock solid at its setpoint.
The question now is whether or not this 20 degree fluctuation is likely to cause analytical difficulties going forward. Any suggestions on the next step?
Thanks for your time,
Kent