Second Calibration not appearing

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello,

I was wondering if anyone has had this happen before?
I have noticed some differences with my injections. Despite the first calibration peaks all showing up, I found that after I ran all my samples, the second calibration peaks did not show up at all. Also, I noticed that despite I do a urea wash after the samples I inject, there is still a huge amount eluted out (which was never the case, usually its very small peaks), almost looking like the sample injection elution. Please help!
Hi,
Quick suggestion
Cause 1: Check your injector and try to prime it 4-5 times
Cause 2: If you using agilent system, Check your Rotor seal if possible change it
Cause 3:If it is an auto sampler, remove your neeedl seat and sonicate with IPA
Is it possible that an overload of samples may cause the calibration peaks to not appear? Since all the sites on the column have been taken up?
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry