Question regarding ECD signal. I installed a refoiled ECD into my 5890 Series II and I get a signal reading of 14574 on the display. It fluctuates but goes back to 14574 if you touch the wire that connects it with the detector board insert. While trying to make sure the detector was seated, a little pressure on the top part of it (that little arm) made the signal go down to <50 (and then taking the pressure off, the signal went back to 14574).
What gives? Do I have a bad ECD? Bad connection?
I found this on another forum:
http://www.labwrench.com/?Home.readPost ... ackground/
What do you guys think?The connectors on the 5890 ECDs are very touchy - you may have a loose or bad connection. If you wiggle it you should see a change in the signal level (actually, just touching it will change it). If you see no change then you probably have a defective electrometer - it is very easy to fry them inadvertently.
John