				contamination
				Posted: Wed Feb 02, 2011 12:09 am
				by natalie
				we perform esterification of fatty acids using acethyl chloride. I am finding contaminating peacks even in a blank sample (with or without addition of the internal standard). Contaminating peaks appear to be stearic and palmitic acids. We performed deep cleaning and maintenance of the instrument. Any ideas?
Thank you.
			 
			
					
				Re: contamination
				Posted: Wed Feb 02, 2011 1:10 am
				by Don_Hilton
				Start with a run with no injection at all, just push the run button, as if you are doing a manual injection.  This will let you know if the problem is in the GC or if it is injected into the GC.
			 
			
					
				Re: contamination
				Posted: Wed Feb 02, 2011 4:01 am
				by chromatographer1
				Please detail the GC and the 'deep cleaning' you mentioned.  Different GCs are prone to different issues.
Rod
			 
			
					
				Re: contamination
				Posted: Wed Feb 02, 2011 7:32 am
				by Peter Apps
				Hi Natalie
 
If these are genuine contaminants, as opposed to carryover, then by far the most likely source is fingerprints on apparatus.  Carryover and contamination can occur together of course.
Peter
			 
			
					
				Re: contamination
				Posted: Wed Feb 02, 2011 4:46 pm
				by natalie
				Thank you for the replies. We think the contamination may derive from the tubes or caps. We use borosilicate glass tubes with polipropylene caps. Do you know any alternative tubes and caps?
Thank you.
			 
			
					
				Re: contamination
				Posted: Wed Feb 02, 2011 5:43 pm
				by Steve Reimer
				Do you have "Hand Soap" in the lab? A common source for these.