Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi, all,

I have an agilent 5975C with MSD (EI and CI) and ECD, I'm wondering if I can cap the ECD when I'm using MSD only for a pretty long time? I don't want the column connected to ECD undergo the oven temperature all the time.

Thank you for any comment :D
If by cap you mean closing the column connection with a nut and blank ferrule then yes, you can. Leave the ECD makeup flowing (you can probably reduce the flow rate a bit) to keep atmospheric contaminants out. Reducing the temperature might prolong foil life by a little bit.

Peter
Peter Apps
Thank you very much, Peter. I'll remove the column and cap the injector and ECD.

If by cap you mean closing the column connection with a nut and blank ferrule then yes, you can. Leave the ECD makeup flowing (you can probably reduce the flow rate a bit) to keep atmospheric contaminants out. Reducing the temperature might prolong foil life by a little bit.

Peter
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry