2-EHA calculation

Discussions about GC and other "gas phase" separation techniques.

9 posts Page 1 of 1
Dear All,

I am doing analysis for 2-Ethyl Hexanoic Acid by GC as BP 2009/EP 7.0. Now the problem for me is the calculation part. In EP/BP the calculation part is given as

[Area of sample x Internal std area in standard x standard weight in g x 2] [/area of std x internal std area in sample x sample weight in g ].

Now my question is what is the role of "2" in the calculation part :?:

Plz explain.........


Regards
Mohammad.
I do not have that monograph at the moment but it is probably dilution of your sample.Probably, it is last extraction volume of your sample.
if you can share your method details , I can answer your question precisely.

Best Regards.
Please note the analysis details:

Inernal Std Preparation: Dissolve 100mg of 3-cyclohexylpropionic acid in 100ml cyclohexane. (1.00mg/ml).

Test solution: 0.3g of substance+4.0ml of conc.Hcl+1ml Internal std. shake vigorously for 1 min. allow the phase to settle. Inject upper layer 1.0µL.

Standard Preparation:Dissolve 75.0mg of 2-EHA in internal std & dilute to 50ml with internal standard.
To 1ml of this solution add 4.0ml con.Hcl. shake vigrously for 1 min, allow the phase to settle. Inject upper layer. (1.0µL).

Capillary Column: Stable wax or DA-FFAP.

Calculation:
(Area ratio of sample/ Area ratio of standard) X (standard wt / sample wt) X 2.
Hi this is Kalidass in the calculation part the 2 is given for dilution of standard and the multiplication of the purity of the 2-EHA. In the method dilution of standard is 0.075g/50 mL. = 100/50=2

Where 100 is the potency of the 2-EHA Standard and 50 is the dilution of the standard 2-EHA solution.

Regards
A.KALIDASS
Regards
A.Kalidass
Thnx Mr.Kalidass for the info....

I would like to ask, instead of taking 1 ml of internal std in std and smaple as describe in the monograph, if i take 3 ml, will there be any change in the results, bcz the concentration of std will be different?

Regards
mohammad.
Hi,

Adding 3 mL of Internal standard in both standard and sample will not differ in the result. Both are resulting is same concentration of 2-EHA. If you want to deviate these things from the monograph then you should validate the same.
I will giving below how the result remains same.

Sample & Standard with 1 mL of internal standard

2-EHA (%) = [(0.075/50)*1/1*1/0.3*100] (Potency of 2-EHA)
= 0.5 %

Sample & Standard with 3 mL of internal standard

2-EHA (%) = 0.075/50*1/3*3/0.3*100 (Potency of 2-ERA)
= 0.5 %

If you adding 3 mL of Internal standard to the sample and 1 mL to standard then the resultant concentration of 2-EHA is now 1.5 %

2-EHA (%) = 0.075/50*1/1*3/0.3*100 (Potency of 2-ERA)
= 1.5 %

Treating the Sample and Standard same will not change in the resultant concentration.
Only the µg/mL will change the µg/g remains as such.

The relation between µg/mL and µg/g is

µg/g= (µg/mL/sample wt in g)* dilution of the sample.

With Best Regards
A.KALIDASS
Regards
A.Kalidass
Once again thnx for clarifying the doubt.....

With respect to test concentration the standard concentration is 0.5% only, where as the limit is 0.8%m/m.
Can you please explain this........... : :?:


Best regards
Mohammad.
Good morning,

I don't know about the specification limit what ever you having. But for the limit 0.8 % we can perform the standard injection at 0.5 %, because if you keeping the limit 0.8% then while doing validation you are going to show the linearity from LOQ Level to 125% or 150% depend on your guideline. Hence we proving our method is linear from LOQ Level to 125% or 150% of specified concentration, we can compare the test solution with any of the standard concentration between LOQ Level to 125% or 150%. It is not compulsory to inject the standard at only specified concentration.
Regards
A.Kalidass
Greetings..

Once again thnx for the info....


Regards
Mohammad.
9 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry