Chlorine in neon or helium

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
I tried searching the topics, but haven't found an answer. I apologize if this has been covered before.
I'm trying to analyze gas mixtures with up to 10% chlorine in helium or neon. We tried a porapak Q, but can't get a reproducible peak. I have also tried neon and helium as carriers, but no luck.
Gow mac 580 GC with TCD;

Any suggestions?
Bubble a known volume through a dilute NaOH solution and then titrate or do a colorimetric test?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
You require gold coated thermisters or wires in your TCD to measure chlorine.

You should use a long copper packed column with a Fluorocol (Krytox 143C) PACKING.

You might wish to use a Carbopack B packing with Cwax phase or a Chromasil 310 or 330 in teflon tubing packed column.

Another assay technique other than GC is recommended.

best wishes,

Rod
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry