Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,

I am actually new to gas chromatography and would like your help in troubleshooting this issue. We have been using a Perkin Elmer Autosystem XL GC using FID technology in our lab to measure the amount of 4-bromostyrene in our samples. We use methanol as the internal standard. The GC has worked fine before but is suddenly not giving any peaks for the sample or the methanol.

I haved tried changing the septa and took a look at the flowrates of the air and hydrogen from the manual as well. Increasing the air flowrates does give peaks but they are also too reduced for any proper results. I would really like your help in determining what could the potential issue be in this regards.

Thanks
Wrong setting on the electrometer of the FID. (computer-integrator)

Leak or hardware problem in the injector that dumps sample other than down the column.

How about a before and after chromatogram to compare if you really need assistance?

good luck,

Rod
Check your syringe.

Check for leaks at inlet.
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry