HP 5890 Win2k ChemStation A.10.01

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello we have an older machine with an Agilent ISA card (model?). I'm trying to configure it to connect to our GC (HP 5890). The software on the machine is Windows 2000 and ChemStation A.10.01. I can upgrade to XP if necessary.

The I/O config tool will recognize the GC address (GBIP 15), but when I try to run ChemStation I get the "Cannot connect to GC, check config" message.

Any thoughts?

Thanks in advance,

-Ian
In the IO config set the SICL interface name to "hp82341" (must be lower case), the bus address to 30 and the system controller box should be checked.
Thanks chromatographer, after doing some more searching on here thats the info that I got as well.

Works great!
Yes, even with an 82350 GPIB board, one types in "hp82341".
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry