HP 5890 Win2k ChemStation A.10.01

Posted: Tue Jan 25, 2011 8:05 pm
by IanH
Hello we have an older machine with an Agilent ISA card (model?). I'm trying to configure it to connect to our GC (HP 5890). The software on the machine is Windows 2000 and ChemStation A.10.01. I can upgrade to XP if necessary.

The I/O config tool will recognize the GC address (GBIP 15), but when I try to run ChemStation I get the "Cannot connect to GC, check config" message.

Any thoughts?

Thanks in advance,

-Ian

Re: HP 5890 Win2k ChemStation A.10.01

Posted: Tue Jan 25, 2011 11:44 pm
by chromatographer
In the IO config set the SICL interface name to "hp82341" (must be lower case), the bus address to 30 and the system controller box should be checked.

Re: HP 5890 Win2k ChemStation A.10.01

Posted: Thu Jan 27, 2011 4:26 pm
by IanH
Thanks chromatographer, after doing some more searching on here thats the info that I got as well.

Works great!

Re: HP 5890 Win2k ChemStation A.10.01

Posted: Thu Jan 27, 2011 6:19 pm
by Consumer Products Guy
Yes, even with an 82350 GPIB board, one types in "hp82341".