water content for biopharmaceuticals

Posted: Mon Jan 24, 2011 5:57 pm
by miro2009
Hello readers,
I was wondering what is the optimum way and device setup to be used for "micro" water content measurement for a lyophilized biopharmaceutical product, that is very hygroscopic? We have no experience with karl fischer titration devices, so are there different models with different sensitivity levels?
We want to use minimum number of pooled vials to avoid excessive exposure to atmosphere during vial opening, emptying and weighing (silica used in weighing area), and our water content per vial shouldn't exceed 0.6 mg, so are there recommended models for "micro-determination" as referred to in the pharmacopea?

Your help and comments are highly appreciated.

Re: water content for biopharmaceuticals

Posted: Mon Jan 24, 2011 8:39 pm
by tom jupille
GC with headspace sampling?

Re: water content for biopharmaceuticals

Posted: Mon Jan 24, 2011 9:19 pm
by krickos
If you refer to like chapter 921 of USP i think you mean something like a coulumeter from for example Metrohm AG.

If you do not have a controlled enviroment around, one might consider placing instrument in glovebox equipped with inert gas stream.

Re: water content for biopharmaceuticals

Posted: Tue Jan 25, 2011 7:04 am
by Markus Laeubli, Metrohm
Dear Miro2009

This really sounds like THE Karl Fischer coulometry application. The coulometric approach allows to analyse very small amounts of water. In your case it should be absolutely no Problem to determine the amount of water in one vial.
Depending of the thermal stability of the substance you add the pharmaceutical to the determination solution or you heat out the water content. There are a lot of applications around.

Do not hesitate to contact your local Metrohm represenative.

Re: water content for biopharmaceuticals

Posted: Sat Jan 29, 2011 1:33 am
by Karen01
This really sounds like THE Karl Fischer coulometry application. The coulometric approach allows to analyse very small amounts of water. In your case it should be absolutely no Problem to determine the amount of water in one vial.
I agree that coulometric Karl Fischer would be fine for residual water post lyo from a single..

BTW I developed a volumetric Karl Fischer assay for lyophilized drug product in a single vial... I could not do coulometric because I had to use a relatively high concentration of chloroform to dissolve the encapsulating microparticle. I used Comp 1 . Handling was a bit tricky to figure out IN THIS CASE but once worked out the assay was robust.

- Karen